Im starting this thread share tips and tricks picked up along the way as well as mistakes and extraction trials. I didn’t want to Hijack / clutter up the MKIII thread so I figured a Terp notebook and pic dump would be a better alternative.....In the shadow of the moon,
Terpene station,
And I know we'll be there soon...
Hopefully, Gray Wolf will be dropping in and answering questions and explaining a few things..
The MKIII operator tech support has been most helpful
GW has done a great job with this unit, he has been awesome at answering any questions I have had,
effectively keeping me on track the whole way.
So here are some pictures of MKIII preparations before the first run…. So here is a picture of the valves. I printed waterproof labels and sealed them on the handles so I could remember which to pull when reading the instructions. At first its kinda hard to remember whats where when you get started. This was great, now I can almost run the machine with my eyes closed. Also included is the operation procedure for the MKIII. Although I think there may be a typo in #15, I assume he meant turn of Vacuum pump instead of the recovery pump as it was turned off in #12...
The Terpenator series are operated as follows:
1.0 Open all valves, except for valve #4, the butane supply valve.
2.0 Turn on high vacuum pump and pump system down to -29.5″Hg.
3.0 Close valves #1 and #3.
4.0 Turn off high vacuum pump and turn on the recovery pump.
5.0 Place one hand on the vent line exiting the top of the column and turn on valve #4.
6.0 When you feel the sudden chill of liquid butane reaching the vent tube, close valve #2, and then valve #4. Record the number of seconds required to flood the column with the existing packing density.
7.0 Continue to pump with the recovery pump, until the internal pressure is reduced to -10″Hg.
8.0 Open valve # 2 and valve # 4 for the same number of seconds that it took to flood the column the first time.
9.0 Close valve #2 and then #4 and continue to pump, until the system pressure is again reduced to -10″Hg.
10.0 Repeat steps #8 and #9 for as many cycles as you choose to run. Typically three cycles for the Mk IA and II, and five cycles for the faster Mk III and IIIA.
11.0 At the end of the last cycle, when the gauge hits -10″Hg, open valve #2 and then valve #3 to dump column.
12.0 Continue to pump until the system reaches -22″Hg, at which point close valve #5 and turn off the recovery pump.
13.0 Open valve #1, and turn on the high vacuum pump.
14.0 Pump the system down to -29.5″Hg and close valve #1.
15.0 Shut off the recovery pump and allow the system to sit under hard vacuum for 5 minutes to purge.
16.0 Remove product from lower collection tank.
Then after connecting a quick release compression fitting into valve 4. Connect to a compressor and begin pressurizing the system to 60 psi first check to see if its holding, then again at 100psi and submege in a barrel of water to check for leaks.
After many attempts to get the MKIII airtight I gnarled the threads to one of the ss connection pieces and had to replace the part. Costs me around 3$ for a SS double ended thread connection. Before the replacement I had a slow ultra tiny leak. Was hard to notice unless you were at the right angle but I managed to snap a pic of the micro bubbles..Highlighted by the red arrow...